i've a mq135 - http://www.futurlec.com/Datasheet/Sensor/MQ-135.pdf

please anyone correct me if i'm wrong:

what i've undestand from datasheet is that

ppm  = Rs/Ro

Rs = mq135 readed resistence
Ro = sensor resistance at 100ppm of NH3 in the clean air (according to datasheet 30Kto200K)

i wire the sensor according to datasheet (see attached image, downloaded is from http://www.staceyk.org)
set RL = pulldown resistor (in the example 39K, according to datasheet any between 10KΩ to 47 KΩ)

so Rs = 1024*RL/adc - RL

the point here is to setup Ro, which means calibrate the sensor.
any help?


thank you,
davide


* MQ135.png (69.05 KB, 278x192 - viewed 53 times.)
Logged

Offline Offline
Edison Member
*
Karma: 9
Posts: 1010
View Profile
 Bigger Bigger  Smaller Smaller  Reset Reset

This is basic information : http://playground.arduino.cc/Main/MQGasSensors
Logged

Global Moderator
Netherlands
Offline Offline
Shannon Member
*****
Karma: 168
Posts: 12425
In theory there is no difference between theory and practice, however in practice there are many...
View Profile
 Bigger Bigger  Smaller Smaller  Reset Reset


MQ sensors were discussed previous year here - http://arduino.cc/forum/index.php/topic,55780.0.html -
Logged

Rob Tillaart

Nederlandse sectie - http://arduino.cc/forum/index.php/board,77.0.html -
(Please do not PM for private consultancy)

0
Offline Offline
Newbie
*
Karma: 0
Posts: 29
Arduino rocks
View Profile
 Bigger Bigger  Smaller Smaller  Reset Reset

yes, i've read the information you link.
but the key point is to find a valid Ro, i do not have a calibrated sensor, so i ask you if someone can tell me a valid Ro values for this chip.

thanks
Logged

Offline Offline
Edison Member
*
Karma: 9
Posts: 1010
View Profile
 Bigger Bigger  Smaller Smaller  Reset Reset

Sorry, but that is not possible. Every sensor could be different.
Logged

0
Offline Offline
Newbie
*
Karma: 0
Posts: 29
Arduino rocks
View Profile
 Bigger Bigger  Smaller Smaller  Reset Reset

i've made a mistake before, i consider Rs/Ro = ppm, but from fig.2 of datasheet is not so.
it seems that it is a power function (see attached).
so i made some power regression based on the values, and i obtain a scaling factor and an exponent
given those values
a = scaling factor obtained by datasheet figure, for the gas you need
b = exponent obtained by datasheet figure, for the gas you need
ppm = a*(Rs/Ro)^b
now, to obtain a valid Ro value, we have to know the ppm of a gas, for C02, we know the CO2 value for atmosphere,
http://www.esrl.noaa.gov/gmd/ccgg/trends/#mlo
so i can obtain Ro value for the chip in open air.

can this be right?


* mq135_test.jpg (58.5 KB, 800x391 - viewed 154 times.)
Logged

Global Moderator
Netherlands
Offline Offline
Shannon Member
*****
Karma: 168
Posts: 12425
In theory there is no difference between theory and practice, however in practice there are many...
View Profile
 Bigger Bigger  Smaller Smaller  Reset Reset

Quote
so i can obtain Ro value for the chip in open air.
can this be right?
Assuming you're not in Bejing with the current smog levels the answer is yes.
Logged

Rob Tillaart

Nederlandse sectie - http://arduino.cc/forum/index.php/board,77.0.html -
(Please do not PM for private consultancy)

0
Offline Offline
Newbie
*
Karma: 0
Posts: 29
Arduino rocks
View Profile
 Bigger Bigger  Smaller Smaller  Reset Reset

Quote
so i can obtain Ro value for the chip in open air.
can this be right?
Assuming you're not in Bejing with the current smog levels the answer is yes.

i hope here near milan is not so high smiley
even if i can know it with precision i will approximate it to 394
Logged

Offline Offline
Newbie
*
Karma: 0
Posts: 2
View Profile
 Bigger Bigger  Smaller Smaller  Reset Reset

But the problem is the MQ-135 sensor is not for CO2. It is for Ammonia, Sulfide, Benzene etc.
I can't calibrate the sensor under the normal atmosphere.
Logged

0
Offline Offline
Newbie
*
Karma: 0
Posts: 29
Arduino rocks
View Profile
 Bigger Bigger  Smaller Smaller  Reset Reset

i'm comparing it against a mh-z14 (using a xively datastrem).
it seams to words good also for Co2 (from 300 to 2000ppm)
results will be published soon on my blog (as soon as i will have time to write the post, and analyze graph).