Arduino Uno I2C HDL3000 PSU
I have access to a HDL3000 PSU and it is connected to an Arduino Uno via I2C.
I found this documentation that explains how to communicate with the HDL3000 from an Arduino using I2C HDS_Series_I2C.pdf .
The code I use:
#include <Wire.h>
const int i2cAdd = 0x50;   // I2C address
 Wire.begin();
  Wire.beginTransmission(i2cAdd);
  Wire.write(0x60);      // Send command
  Wire.endTransmission(false);
  Wire.requestFrom(i2cAdd,2);
  while(Wire.available()){
    char c = Wire.read();
    Serial.print(c);
  delay (100);
When I send the command "0x00" to get "Unit Manufacturer" the response is "XP Power Ltd".
But when I send the command "0x60" to get "Output Voltage Measurement" the response is empty.
The same result is for "Output Current Measurement", "Output Voltage Setting" and "Output Current Setting".
I don't understand why.
Any idea?
Is this the correct command? Shouldn't the argument be "true"?
No, the datasheet shows no stop condition after the write.
What does empty mean? The result won't be in ASCII, so you might have to change the code line
to
Serial.print((uint8_t) c, HEX);
to get a readable version of the return value.
Great!
I tried to read "Output Voltage Measurement" from the PSU.
  Wire.requestFrom(i2cAdd,1);
I got "0xFF"
And then:
  Wire.write(0x61);      // Send command
I got "0xB"
But this doesn't sound right.
In the documentation it states that to get "Output Voltage Measurement" the following command is sent to the PSU:
[S_][A0][60][RS][A1][02][P_]
I found some information on this document : DV164122_serial_analyzer_eng_users.pdf
It states on page 36:
• Start bit (S_)
• Slave Address[W] – Enter the slave address of the device to communicate with.
The Write bit should be cleared to indicate a write operation.
• Word Address – Enter the word address
• Restart (RS),
• Slave Address[R] – The slave address with the Write bit set will be automatically
entered when the Slave Address[W] has been entered.
• Byte Count – Enter the number of bytes to be read
• Stop bit (P_)
How to translate this command in Arduino code?
That means the device measures a voltage of 0x0BFF cV = 30.71V. Does that make sense?
The second half of page 5 explains how to interpret the bytes.
You almost had it:
  uint16_t l = Wire.read();
  uint16_t h = Wire.read();
  uint16_t v = h << 8 | l;
  float voltage = (float) v / 100.0;
When I try your code I get 655.35.
l = 255 (0xFF)
h = 255 (0xFF)
v = 65535 (0xFFFF)
The voltmeter shows cV=29.90V.
This is the more explicit version. If that doesn't work, post a complete wiring diagram and a sharp picture of your setup.
Your code works as expected.
It shows 30.71
But when I check with the voltmeter I get 29.90V
A difference of 0.81V.
I believe that the code is correct.
I will try to set another voltage value and then read the it.
Anyway I appreciate your help.
Thanks!
Is that a high quality meter? According to the datasheet the output voltage has an accuracy of +/-1%, so about a half of the difference might be on the PSU side, cheap voltage meters often have an accuracy in the same range.
In fact I tested with a Fluke 115 and also with a UNI-T UT33A and both show 29.9 V.
I don't understand why it is so.
I am trying to sett the voltage according to the document on page 6.
First we send [S_][50][70][E8][03][P_] and then [S_][50][7C][85][P_]
My code looks like this:
  // Write command
  // Output Voltage Setting
  // 10V
  Wire.write(0x70);
  Wire.write(0xE8);
  Wire.write(0x03);
  // Control Register Setting
  Wire.write(0x7C);
  Wire.write(0x85);
On the PSU the status LED turns Solid (Orange) meaning status "DC Output OK in remote control mode".
But when I read the output voltage I get this:
25.59
0.11
0.00
0.00
0.00
0.00
I use the same code as earlier to read the voltage.
Apparently the code to set the voltage is not correct.
Here the false parameter is wrong, it needs a stop condition. So either set it to true or don't set it (to use the default which is true).
The same result!
Here is the updated code:
  // Write command
  // 10.00V: 1000 = 0x03E8
  // Write 0xE8 to address 0x70
  // Write 0x03 to address 0x71
  Wire.write(0x70);
  Wire.write(0xE8);
  Wire.write(0x03);
  Wire.endTransmission();
  // Write 0x85 to address 0x7C
  Wire.write(0x7C);
  Wire.write(0x85);
Did i miss something?
Maybe a consecutive write is also not possible (as consecutive reads), so you might have to do three writes:
// Write command
  Wire.write(0x70);
  Wire.write(0xE8);
  Wire.write(0x71);
  Wire.write(0x03);
  Wire.write(0x7C);
  Wire.write(0x85);
No, but a full sketch would be nice.
Are you talking to I2C address 0x50 ? Did you run a I2C Scanner sketch ?
I assume that multiple data bytes can be read and written, but pylon is right, it is somewhat vague.
There are two registers that are 16-bit and also Read/Write. You can do a test if you can write two bytes and read two bytes and then check with other values.
Here is the entire code:
#include <Wire.h>
  Wire.begin();
  setOutputVoltage();
  readOutputVoltage();
void readOutputVoltage(){
  // Read command
  // Read Output Voltage Measurement
  Serial.print(voltage);
void setOutputVoltage(){
  // Write command
  // 10V: 1000 = 0x03E8
  Wire.write(0x70);
  Wire.write(0xE8);
  Wire.write(0x71);
  Wire.write(0x03);
  Wire.write(0x7C);
  Wire.write(0x85);
After a couple of readings I get this:
30.71
25.59
0.15
0.00
0.00
0.00
How do you get a couple of readings ? Do you reset the Arduino board ?
It might need a delay between I2C sessions.
Can you add a delay of 1ms after each I2C session ?
You could try a lower or higher clock, for example 50kHz and 200kHz
Wire.begin();
Wire.setClock(50000UL);       // 50kHz as a Unsigned Long
You could test the error returned by Wire.endTransmission() and the number by Wire.requestFrom();
  int error = Wire.endTransmission(false);
  if(error != 0)
    Serial.println("O no, communication lost with the sensor");
  int n = Wire.requestFrom(i2cAdd,1);
  if(n == 1)         // same amount as requested ?
    Serial.println("Something wrong, communication lost with the sensor");
How is your bus ? What are your pullup resistors ? What kind of cable or wires do you use ?
Yes I reset the Arduino board after that I commented out setOutputVoltage();
Yes I did, no difference.
Yes I did, no error.
As I know the Uno has pullup resistors built in that pull the SDA and SCK to 5 volts.
Wiring:
#include <Wire.h>
  Wire.begin();
  //Wire.setClock(100000);
  //Wire.setClock(200000UL);
  setOutputVoltage();
  // Read command
  int result1 = Wire.endTransmission(false);
  delay(10);    // Delay 10ms
  Serial.print("Result1: ");
  Serial.print(result1);
  int result2 = Wire.endTransmission(false);
  Serial.print("Result2: ");
  Serial.print(result2);
  int result3 = Wire.requestFrom(i2cAdd,1);
  Serial.print("Result3: ");
  Serial.print(result3);
  // Write command
  Wire.write(0x70);
  Wire.write(0xE8);
  Wire.write(0x71);
  Wire.write(0x03);
  Wire.write(0x7C);
  Wire.write(0x85);
Output:
Result1: 0
Result2: 0
Result3: 1
25.19
Result1: 0
Result2: 0
Result3: 1
1.26
Result1: 0
Result2: 0
Result3: 1
0.03
Result1: 0
Result2: 0
Result3: 1
0.00
It seems that the communication is okay, the device is really returning zeros.
Result1 and 2 are the error codes and are zero, that is okay.
Result3 is the return value of Wire.requestFrom(), that is also okay since 1 byte was requested, so it should return 1.
Somehow the device gets tired of sending data.
I did not read everything, can you find example of others ?
Can you put it in the loop with a 5 second delay at the end of the loop() ?
Maybe try an other register, maybe include reading the manufacturer string, so you know that everything is working.
They say that a pullup resistors of 2k are needed.
I also added pullup resistors 2k with no change.
I ordered the same interface equipment that is used in the document, PICkit serial analyzer.
I am wandering if this PSU really works.
Hello,
How did it end up ? were you able to program the power supply ? In my case, I am trying to use Rasberry Pi I2C to communicate and still no luck. I don't detect any device at all. Should the wiring be the same ?