CCS811 TVOC sensor issue(s)

Hi,
I'm trying to use a CCS811 sensor module to measure VOC's emitted from a device which momentarily melts small pieces of soft PVC tubing. I have the module connected exactly as shown in the attached image. I am using the example sketch from the adafruit CCS811 library as follows...


Adafruit_CCS811 ccs;

void setup() {
  Serial.begin(9600);

  Serial.println("CCS811 test");

  if(!ccs.begin()){
    Serial.println("Failed to start sensor! Please check your wiring.");
    while(1);
  }

  // Wait for the sensor to be ready
  while(!ccs.available());
}

void loop() {
  if(ccs.available()){
    if(!ccs.readData()){
      Serial.print("CO2: ");
      Serial.print(ccs.geteCO2());
      Serial.print("ppm, TVOC: ");
      Serial.println(ccs.getTVOC());
    }
    else{
      Serial.println("ERROR!");
      while(1);
    }
  }
  delay(500);
}

my question is regarding the delay function... for some reason I am getting readouts approximately 1 per second. even though the delay is set to 500ms. even if i lower the delay to 100ms it still only prints a reading once every second. Can anybody shed some light on why this might be please?

Thanks
this is the sensor I'm using: (CJMCU CCS811 811 Gas Sensors Air Quality Sensor Module for ESP8266 ESP32 Arduino | eBay)

arduino-sensor-c02-ccs811-conexion

Datasheet sez:

The CCS811 supports
multiple drive modes to take a measurement every 1 second, every 10 seconds, every
60 seconds, or every 250 milliseconds.

It also mentions it does I2C "clock stretching" that can slow things down.

Modes of Operation
The CCS811 has 5 modes of operation as follows
• Mode 0: Idle, low current mode
• Mode 1: Constant power mode, IAQ measurement every
second
• Mode 2: Pulse heating mode IAQ measurement every 10
seconds
• Mode 3: Low power pulse heating mode IAQ
measurement every 60 seconds
• Mode 4: Constant power mode, sensor measurement
every 250ms
In Modes 1, 2, 3, the equivalent CO2 concentration (ppm) and
TVOC concentration (ppb) are calculated for every sample.
• Mode 1 reacts fastest to gas presence, but has a higher
operating current
• Mode 3 reacts more slowly to gas presence but has the
lowest average operating current.
When a sensor operating mode is changed to a new mode with
a lower sample rate (e.g. from Mode 1 to Mode 3), it should be
placed in Mode 0 (Idle) for at least 10 minutes before enabling
the new mode. When a sensor operating mode is changed to a
new mode with a higher sample rate (e.g. from Mode 3 to Mode
1), there is no requirement to wait before enabling the new
mode.
Mode 4 is intended for systems where an external host system
wants to run an algorithm with raw data and this mode provides
new sample data every 250ms.