Need guidance in photodetection part of DAQ system

I am currently working on a project to develop a fibre optic dosimetry system. The system uses an organic scintillator (BCF-60), which emits green light at 530 nm and has an active volume of 0.8 mm³.

As part of the project, I am designing a transimpedance amplifier (TIA) circuit using the S1223-01 photodetector, which I selected based on its spectral response range (360 nm to 900 nm) and photosensitivity (0.1–0.6 A/W).

To simulate a clinical environment, I use a function generator to produce a 360 Hz pulse signal, which drives a 5 mm green LED. The light signal is transmitted through a 2-meter-long, 1 mm diameter fibre optic cable. The other end of the fibre is coupled to the photodiode, which is connected to the TIA circuit built around the OP07CP op-amp, using a feedback resistor of 1.1 MΩ. The output of the TIA is then connected to a second-stage amplifier (a voltage follower using the OP276) configured with unity gain.

However, I am currently experiencing noise in the output signal, and I plan to test the setup using a National Instruments DAQ (NI 6363) for further analysis.

I attached circuit diagram in which I am using TIA (OP07CP) I am getting noise of around 100 mV at output of TIA. kindly please guide me how I could reduce this noise.

Hasham

The starting point for noise investigation is to measure the noise signal at the photodiode/transimpedance amp output alone, with the photodiode and amp in complete darkness.

Your circuit shows no details of the op amp power supplies and no decoupling caps on any of the ICs. Those are absolutely critical design details for successful amplification of low level signals.

Please post a photo of your setup. If you have this circuit lashed together with looping wires on a hobby breadboard, start over.