Hey guys,
Ive built a non invasive ventilator and it uses 2 pressure sensors to measure the pressure of the exit valve of the ventilator which leads to the mask. This is a project im doing for school and its due very soon and i need this to be working so any help would be appreciated.
The issue on startup the pressure sensors are calibrated, sometimes the works and the pressure sensors are zeroed but one of them sometimes does not calibrate and has random fluctuating output. This then stops my ventilator from functioning as it does not start until they are calibrated.
here is the schematic im using:
Here is my circuit diagram for the main board:
Here is the circuit diagram for the pressure sensor board which is connected to the main board via jumpers:
NOTE: I have removed the 5k and 5ohm resistors and the diodes this was a suggestion from someone on another forum
here is the code im using:
//Source code for the BIPAP/CPAP low cost device v1.0
//Arduino Nano
//Details of the hardware, connections and setup can be found in www.ub.edu/biofisica
//For code details contact jorge.otero@ub.edu / m.rodriguez@ub.edu
#include <PID_v1.h> //PID library
double Setpoint, Input, Output, flow_hp, flow_bp; //variables
double Kp=0.38, Ki=0.75, Kd=0; //variables and values for the PID controller KP(proportional),Ki(integral),Kd(derivative)
//Definitions
PID myPID(&Input, &Output, &Setpoint, Kp, Ki, Kd, DIRECT);
#define analog 10 //define digital 10 as analog.
#define analog_flow 9 //define digital 9 as analog_flow.
#include <LiquidCrystal.h> //LCD 2x16
LiquidCrystal lcd(12,11,5,4,3,2); //LCD connections
#define switchPin 6 //define digital 6 as SwitchPin.
#define SwitchMode 13 //define digital 13 as SwitchMode.
#define led 8 //define digital 8 as led.
byte autoipap = 0,lcdmodecpap=1,lcdmodebipap=1,switchState = 0,switchModestate = 0,restard; //byte variables
unsigned long t0 =micros(),tiempoahora =millis();
int Tm =10000; //Sampling frequency in microseconds
float tiempo =0;
float fc_hp=0.125; //cutoff frequency of the high-pass filter
float fc_lp=1.125; //cutoff frequency of the low-pass filter
float RC_hp =1/(2*PI*fc_hp);
float RC_lp =1/(2*PI*fc_lp);
double data_filt_hp[] = {0, 0};
double data_hp[] = {0, 0};
double data_filt_lp[] = {0, 0};
double data_lp[] = {0, 0};
double maximo=0,vmax=0,flow=0;
int pressure, epap,ipap,consigna=0;
float porcentaje,P_sensor,P_epap,P_ipap,P_sensor_abs,P_sensor2;
int ctroltime =0;
int gain_flow =100;
char P_sensorstring[4];
char P_sensorstring2[4];
char P_EPAPstring[4];
char P_IPAPstring[4];
char porcentajestring[2];
char timestring[2];
int i,z=0;
int samplepressure[25]; //buffer of 25 positions for the pressure sensor 1 readout
int samplepressure2[25]; //buffer of 25 positions for the pressure sensor 2 readout
double totalpress2 = 0.00, promediopress2 = 0.00, totalpress = 0.00, promediopress = 0.00;
double totalzero=0,totalzero2=0,samplezero,samplezero2,pressinzero,pressinzero2,zero,zero2;
#define sensorpress A0 //pressure sensor 1 output is connected to the analog input 0
#define sensorpress2 A1 //pressure sensor 2 output is connected to the analog input 1
#define analog_epap A2 //epap potentiometer adjustment is connected to the analog input 2
#define analog_ipap A3 //ipap potentiometer adjustment is connected to the analog input 3
#define analog_porcentaje A4 //cycling potentiometer adjustment is connected to the analog input 4
//Configuration
myPID.SetMode(AUTOMATIC); //automatic mode in PID
consigna=ipap;
for(i=0; i< 24; i++) //array initialization
samplepressure[i] = 0;
samplepressure2[i] = 0;
i=0;
lcd.begin(16,2); //LCD 2x16
pinMode(switchPin,INPUT); //switchPin -> input
pinMode(SwitchMode,INPUT); //switchMode -> input
pinMode(led,OUTPUT); //led -> output
lcd.setCursor(0,0);
lcd.print("Calibrating zero");
lcd.setCursor(0,1);
for(i=0; i< 91; i++) //array initialization
Serial.println(i);
if (i==0 || i==6 || i==12 || i==18 || i==24 || i==30 || i==36 || i==42 || i==48 || i==54 || i==60 || i==66 || i==72 || i==78 || i==84 || i==90)
lcd.setCursor(i/6,1);
lcd.print(".");
samplezero = 2.5*analogRead(sensorpress); //zero reading for pressure sensor 1
totalzero = totalzero + samplezero;
samplezero2 = 2.5*analogRead(sensorpress2); //zero reading for pressure sensor 2
totalzero2 = totalzero2 + samplezero2;
delay(50);
zero=totalzero/i;
zero2=totalzero2/i;
lcd.setCursor(0,1);
lcd.print(" ");
lcd.setCursor(0,1);
lcd.print("P1:");
lcd.setCursor(8,1);
lcd.print("P2:");
lcd.setCursor(15,1);
lcd.print(" ");
for(i=0; i< 50; i++)
pressinzero =2.5*analogRead(sensorpress)-zero; //zero adjustment for pressure sensor 1
P_sensor=abs((pressinzero/1023)*5.0*4.8);
dtostrf(P_sensor,1,1, P_sensorstring);
pressinzero2 =2.5*analogRead(sensorpress2)-zero2; //zero adjustment for pressure sensor 2
P_sensor2=abs((pressinzero2/1023)*5.0*4.8);
dtostrf(P_sensor2,1,1, P_sensorstring2);
lcd.setCursor(4,1);
lcd.print(P_sensorstring);
lcd.setCursor(12,1);
lcd.print(P_sensorstring2);
delay(30);
if (P_sensor >0.3 ||P_sensor2>0.3)
lcd.setCursor(0,0);
lcd.print("err calibrating ");
lcd.setCursor(0,1);
lcd.print(" ");
lcd.setCursor(4,1);
lcd.setCursor(0,1);
delay(2000);
lcd.print("Restart sofware.");
restard =1;
i=50;
i=0;
// Main program loop
if (restard ==1)
i=0;
totalzero=0;
totalzero2=0;
restard=0;
setup();
samplepressure[i] = 2.5*analogRead(sensorpress); //Pressure sensor 1 sampling
totalpress = totalpress + samplepressure[i];
samplepressure2[i] = 2.5*analogRead(sensorpress2); //Pressure sensor 2 sampling
totalpress2 = totalpress2 + samplepressure2[i];
ctroltime = 5;
i++;
if (micros()-(t0)>=Tm)
tiempo=(micros()-t0);
t0=micros();
promediopress = (totalpress/i); //pressure average for sensor 1
promediopress2 = (totalpress2/i); //pressure average for sensor 1
pressinzero = promediopress-zero; // zero adjustment for pressure sensor 1 - sensor should be let with no pressure and the zero adjusted with the potentiometer
pressinzero2 = promediopress2-zero2; // zero adjustment for pressure sensor 2 - sensor should be let with no pressure and the zero adjusted with the potentiometer
P_sensor=(pressinzero2/1023)*5.0*4.8; //conversion to pressure in cmH2O units.
P_sensor_abs = abs(P_sensor);
Serial.println(P_sensor_abs);
epap=0.833*analogRead(analog_epap)+170; //epap range from 4 cmH2O to 24 cmH2O (170-1023)
ipap= 0.833*analogRead(analog_ipap)+170; //ipap range from 4 cmH2O to 24 cmH2O (170-1023)
if (epap>ipap)
epap=ipap; //epap value never can be higher than ipap value
P_epap =(epap/1023.00)*5.0*4.8; //conversion of epap pressure to cmH2O units
P_ipap =(ipap/1023.00)*5.0*4.8; //conversion of ipap pressure to cmH2O units
porcentaje=100*(analogRead(analog_porcentaje)/1023.00); //conversion of the cycling value to percentage (0% - 100%)
flow=gain_flow*(pressinzero2-pressinzero); // flow calculation from the pressure difference in the sensors
if (flow>0)
flow=100*sqrt(flow); //flow+ linearization
flow=-100*sqrt(-flow); //flow- linearization
int flow_out = map(flow,-20000,40000,0,255); // rescaling flow values (0-255) for PWM control in pin 9
analogWrite(analog_flow,flow_out);
Input =pressinzero2; //pressure of the transducer is the input for the PID controller
Setpoint= consigna; //setpoint for the PID controller
myPID.Compute(); //myPID function call
if (pressinzero2 < 250)
Output=85; //if pressure is very low it is because there is no patient, so output is settled very low
analogWrite(analog,Output); //analog output 10 actuates the blower
switchState = digitalRead(switchPin); //adjust values switch readout
switchModestate = digitalRead(SwitchMode); //CPAP/BIPAP modes switch readout
if (switchModestate == LOW && switchState == LOW) //CPAP mode is active when both switches are off
if (lcdmodecpap ==1)
myPID.SetTunings(Kp=0.12, Ki=0.75, Kd=0); //PID values for CPAP mode
lcd.setCursor(0,0);
lcd.print(" CPAPmode ");
lcd.setCursor(0,1);
lcd.print("Press ");
lcd.setCursor(11,1);
lcd.print("CmH20");
lcdmodecpap = 0; // no rewritting of the display until mode is switched again
digitalWrite(led , LOW);
lcd.setCursor(7,1);
dtostrf(P_sensor_abs,2,0, P_sensorstring);
lcd.print(String(P_sensorstring));
consigna=ipap; //setpoint is adjusted by the ipap potentiometer
lcdmodebipap=1; //allows switching to BIPAP mode
if (switchModestate == LOW && switchState == HIGH) //CPAP mode, adjustment of the parameters
lcd.setCursor(0,0);
lcd.print(" Set CPAP ");
lcd.setCursor(0,1);
dtostrf(P_ipap,2,1, P_IPAPstring);
lcd.print(String("IPAP " )+String(P_IPAPstring)+String(" CmH2O ")); //display the IPAP value
lcdmodebipap=1; //allows mode switching
lcdmodecpap=1; //allows mode switching
if ( switchModestate == HIGH && switchState == LOW) //BIPAP mode is active, no parameters setup
if (lcdmodebipap ==1)
lcd.setCursor(0,0);
lcd.print("BiPAPmode ");
lcd.setCursor(12,0);
lcd.print("%");
lcd.setCursor(0,1);
lcd.print("CmH20");
lcdmodebipap=0; // LCD is just written once
dtostrf(porcentaje,2,0,porcentajestring); //conversion of % to string
lcd.setCursor(10,0);
lcd.print(String(porcentajestring));
lcd.setCursor(7,1);
dtostrf(P_sensor_abs,2,0, P_sensorstring); //conversion of pressure to string
Serial.println(P_sensorstring);
if (autoipap==0)
lcd.setCursor(13,0);
dtostrf(ctroltime,1,0,timestring); //conversion of time to string
lcd.print(String(" -")+String(timestring));
if (switchModestate == HIGH && switchState == HIGH) //BIPAP mode, parameter adjustment
lcd.setCursor(0,0);
lcd.print(String("IPAP ")+String(P_IPAPstring)+String(" CmH2O "));
lcd.setCursor(0,1);
dtostrf(P_epap,2,1, P_EPAPstring);
lcd.print(String("EPAP ")+String(P_EPAPstring)+String(" CmH2O "));
lcdmodebipap=1;
lcdmodecpap=1;
flow_hp = mifiltroHP(flow,tiempo/1000000,RC_hp); //high-pass filtering flow signal
flow_bp = mifiltroLP(flow_hp,tiempo/1000000,RC_lp); //low-pass filtering flow signal
if (flow_bp>350 && switchModestate == HIGH && switchState == LOW) //detection of inspiratory flow > 350 when BIPAP mode is active
myPID.SetTunings(Kp=0.1, Ki=0.45, Kd=0); //new PID parameters
digitalWrite(led , HIGH); //inspiration detection LED is active
tiempoahora=millis();
maximo = max(maximo, flow_bp); //maximum flow calculated
if (maximo>1.05*flow_bp)
{ //detects a 5% flow reduction respect to the peak flow
vmax =porcentaje/100*maximo; //calculation of the change point to EPAP (Vmax)
if (flow_bp<vmax)
consigna = epap; //if the flow is lower than the change point, switches to EPAP mode
myPID.SetTunings(Kp=0.65, Ki=0.2, Kd=0); //new PID parameters
consigna=ipap; //if the flow is inspiratory but not 5% reduced, it maintains IPAP mode
if (flow_bp<350 && switchModestate == HIGH && switchState == LOW) //detect flow < 350 in BIPAP mode
digitalWrite(led , LOW); //switch off inspiration led (flow<350 = expiration)
if (millis()>(tiempoahora+(1000*ctroltime))&& millis() <= (tiempoahora+(1000*ctroltime)+1010)) //detects if the flow<350 time is greater than control time
autoipap=1; //autoIPAP mode activated
consigna=ipap; //IPAP mode activated
lcd.setCursor(14,0);
lcd.print(String(F("ON")));
if (millis()> tiempoahora+(1000*ctroltime)+1000)
consigna=epap;
lcd.print(String(timestring));
{ //no autoIPAP mode needed
consigna=epap;
autoipap=0;
if (millis()> tiempoahora+(1000*ctroltime)+3000)
tiempoahora=millis(); //time variable reset
maximo=0; //maximum varibles reset
vmax=0;
if (switchModestate == LOW && switchState == LOW) //CPAP mode active, maintains IPAP
consigna=ipap;
if (switchModestate == LOW && switchState == HIGH) //CPAP mode active, setting parameters, maintains IPAP
consigna=ipap;
//variables reset
promediopress=0;
totalpress=0;
promediopress2=0;
totalpress2=0;
i=0;
} //end of main program loop
double mifiltroHP(double dato,double dt,double RC) //high-pass filtering of flow function
double alpha = RC/(RC+dt);
data_hp[1] = dato;
// High Pass Filter
data_filt_hp[1] = alpha * (data_filt_hp[0] + data_hp[1] - data_hp[0]);
// Store the previous data in correct index
data_hp[0] = data_hp[1];
data_filt_hp[0] = data_filt_hp[1];
return (data_filt_hp[1]);
double mifiltroLP(double dato,double dt,double RC) //low-pass filtering of flow function
double alpha = dt/(RC+dt);
data_lp[1] = dato;
// low Pass Filter
data_filt_lp[1] = alpha *data_lp[1]+(data_filt_lp[0]*(1-alpha));
data_lp[0] = data_lp[1];
data_filt_lp[0] = data_filt_lp[1];
return (data_filt_lp[1]);
Here is what i see on my LCD when it doesnt seem to want to calibrate:
I ran this code to test the output of the A1 pressure sensor the one that seems to give a fluctuating value at times:
int Rawval; // ADC value
int Pressoffset = 102; // ADC value @ 0 pressure
int adcMax = 921; // ADC value @ 5kPa (51 cmH2O)
int span;
int Newval;
span = adcMax - Pressoffset; // 819 here
Rawval = analogRead(A0);
Serial.print(" Rawval = \t");
Serial.print(Rawval);
Newval = Rawval - Pressoffset;
Serial.print("\t Pressure = \t");
float pressureVal = (float)Newval * 51.0 / (float)span;
Serial.print(pressureVal, 1); // cmH2O
Serial.println(" cmH2O");
delay(2000);
Here are the results: